Pediatric Bone Marrow Transplant: A Guide for Families

Emergency Procedure For A Broken Or Leaking Central Line Catheter [Home] - PBMT

Geri Quinn, RN, MSN and Janine Petitgout, RN, MA
Peer Review Status: Internally Peer Reviewed


Equipment:

  • Alcohol swabs
  • Sterile scissors
  • Blunt needle
  • Pediatric Broviac® catheter...20 gauge
  • Adult Broviac® catheter...18 gauge
  • Hickman® catheter...15 gauge
  • Double lumen:
    • 7 Fr. small lumen...20 gauge
    • 7 Fr. large lumen...18 gauge
    • 9 Fr. both lumen...18 gauge
    • 12 Fr. small lumen...18 gauge
    • 12 Fr. large lumen...15 gauge
  • Sterile luerlock catheter plug
  • Tongue blade
  • Preservative heparin (10 u/cc)
  • Clamp
  • Sterile luerlock syringes with needles (2)
  • 4 x 4 topper
  • Tape

Procedure:

1. If the catheter breaks, leaks or cracks, it should be immediately clamped between the child and the damaged portion of the catheter.

BMT

2. Wash hands thoroughly.

BMT

3. Wipe top of heparin vial with alcohol and draw up 3 cc heparin.

BMT

4. Change needle on syringe to appropriate size blunt needle.
5. Wipe damaged portion of catheter with alcohol swab.

6. Trim off damaged portion of catheter with sterile scissors.

BMT

7. Insert blunt needle into internal sheath of catheter. This will form a temporary luer stub adapter.

BMT

8. Open clamp and irrigate with heparin, clamping catheter as last of heparin is instilled.

BMT

9. Remove the syringe, leaving the needle in the catheter, and attach a new catheter plug.

BMT

10. Tape repaired section of catheter to tongue blade padded with 4 x 4 topper. This will stabilize the repaired section of the line so that the blunt needle cannot puncture the catheter.

BMT

11. Flush other port with heparin.

*Temporary repairs can only be done on single lumen and double lumens that break before the bifurcation. If you have a double lumen that breaks above the bifurcation (where one lumen is divided into two), you cannot do a temporary repair. Just clamp the line and call your physician. You will need to come in for a permanent repair after doing the temporary repair.


Injection of Medications Through Central Lines - PBMT

Equipment:

  • Alcohol swabs
  • Syringes with 3 cc normal saline (2)
  • Syringe with medication
  • Syringe with 3 cc (10 u/cc) preservative heparin
  • Sterile luerlock catheter cap
  • Tape

Procedure:

1. Wash hands.

BMT

2. Clean catheter connection with alcohol.

BMT

3. Clamp catheter.
4. Remove old cap.

5. Connect syringe with normal saline. Unclamp. Instill saline. Clamp. Remove syringe.

BMT

6. Connect syringe with medication. Unclamp. Instill medication over designated time. Clamp catheter. Remove syringe.
7. Connect syringe with normal saline. Unclamp. Instill saline. Clamp. Remove syringe.
8. Connect syringe with heparin. Unclamp. Inject heparin, clamping catheter as last of 3 cc is instilled.
9. Remove syringe and wipe catheter with alcohol swab.

10. Attach new cap.

BMT

11. Tape catheter cap in place.

Title Page



Department of Pediatrics